Air tightness test failed

By
Melanie Schwabe
on 9/13/18, 9:40 AM 92 views

After gripping the probe, the automated air tightness test failed or gave a warning. What shall I do now?

This question has been included in the official documentation.

0

Melanie Schwabe

--Melanie Schwabe--

5355
| 3 1 2
Glattbrugg, Switzerland
--Melanie Schwabe--


Melanie Schwabe
On 9/13/18, 9:51 AM

The leak flow is indicated in Pascal [Pa] because any leak flow generates a pressure drop along the pneumatic tubing, which we can measure with our pressure sensors. In a perfectly tight system the air flow should be 0 ul/s and the corresponding pressure drop is also 0 Pa.

Warning < 1 Pa = no problem

Values below 1 Pa are still safe to work, despite the warning by the system. Continue with your experiment as usual.

> 1Pa = something is wrong

In case the leak generates a pressure drop above 1 Pa, troubleshooting is required:

  1. Gripping: Drop and grip the probe again. Often this resolves the issue.

  2. Gripping height calibration: If you can see that gripping looks "odd" and the probe does not come into the center, please calibrate the Z stage level again. Also calibrate the XY position of the Z stage again.

  3. Tubing: Check if the tubing is attached properly and check if there are visible cracks, holes etc in the tubing

  4. Dirty head: The sealing screw and/or prism are dirty or the O-ring is damaged. Please follow the instructions here

  5. Probe defect: In rare cases the consumable might have a defect if the above methods didn’t solve the problem. Have a look under the stereoscope onto the probe reservoir if you see any cracks or chipped parts please contact your FluidFM distributor sending a detailed written description of the problem.

  • Flag

Your Answer

Please try to give a substantial answer. If you wanted to comment on the question or answer, just use the commenting tool. Please remember that you can always revise your answers - no need to answer the same question twice. Also, please don't forget to vote - it really helps to select the best questions and answers!

Keep Informed

About This Community

This community is for professionals and enthusiasts of our products and services. Share and discuss the best applications, experiments and new ideas, build your professional profile and become a better researcher together. Read Guidelines

Question tools

1 follower(s)

Stats

Asked: 9/13/18, 9:40 AM
Seen: 92 times
Last updated: 10/12/18, 9:58 AM